Re: cell count

From: Dr S Sreckovic (Sasa.Sreckovic@bristol.ac.uk)
Date: Tue Dec 07 1999 - 10:52:01 EST


Hi Ann,

I just want to express slight disagreement on sorting counters. 
On BD machines SORT L and R counters are actually counting deflection 
pulses, meaning that number given is:

events that fit gates - (analysis aborts + coincidence aborts).

Analysis aborts can be caused in three ways. Most important type is 
threshold abort - if second event occurs within the dead time of the 
previous one or if the first pulse exceeds its own dead time interval 
(pulse fails to drop bellow threshold by its end).
On the other hand coincidence aborts would depend on sort mode, drop 
envelope size and drops/1cell ratio (drive frequency/event rate).

Cheers,
Sasha.

On Mon, 06 Dec 1999 12:02:07 +0100 Ann Atzberger 
<Ann.Atzberger@EMBL-Heidelberg.de> wrote:

> 
> 
> Hi Chris,
> 
> in our case the cells were actually lysing as they weren't compatible with
> the medium being used. 
> I'd also do a drop delay check. The cell number on the instrument counts
> all the cells that fit in the sort gates but that does not mean it has
> actually sorted it.
> 
> regards
> Ann
> 
> At 13:35 02.12.99 -0800, you wrote:
> >
> >Hello gang,
> >My question deals with our FacStar. I am getting reports of counts that are
> >about 1/2 of the cell number the instrument says were recovered. I am
> >wondering if anyone has seen this or what to do to fix this problem. Cell
> >count is very important in many of our studies.
> >Thanks in advance and Happy Holidays,
> >
> >chris
> >
> >
> >
> >
> >
> >Chris Lena
> >Laboratory Technician
> >La Jolla Institute For Allergy and Immunology
> >phone:(858)678-4536
> >fax:(858)678-4595
> >cali@liai.org
> >http://www.liai.org
> >
> >               _   _
> >              (_)-(_)
> >                \"/
> >                =V=
> >
> >            
> >
> >
> >
> 

----------------------
Dr Sasha Sreckovic
Dept Path & Micro
University of Bristol
University Walk
Bristol, BS8 1TD, UK
Sasa.Sreckovic@bristol.ac.uk
0117-928-8606



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:54:18 EST