Re: FW: Compensation for Green contaminating RED channel

From: Howard Shapiro (hms@shapirolab.com)
Date: Wed Dec 01 1999 - 23:19:54 EST


Regarding Jim Leary's suggestions about compensation -

Mark Corio's 3-color compensation electronics are indeed quite 
nice.  However, as one adds colors, particularly when the labels used are 
tandems in which energy transfer is incomplete, or when there is energy 
transfer between dyes, as can occur with fluorescein-labeled antibodies to 
nuclear antigens and 7-AAD on nuclear DNA, it becomes necessary to have 
nonzero values for most elements of the compensation matrix, meaning more 
circuitry if compensation is done in hardware.  This circuitry must be 
placed  between the PMT preamp outputs and the log amp inputs, and the 
analog electronics needed to compensate for more than four colors, even 
when designed and built with a great deal of both skill and care, are 
almost certain to raise the noise level to the point at which dynamic range 
is compromised.  The problem is avoided by collecting linear data with 16 
or more bits' resolution, and doing both compensation and log 
transformation with digital computers, as is done in the Beckman Coulter 
Epics XL and a number of laboratory-built instruments; systems now in 
development incorporate digital pulse processing, which further improves 
performance, particularly at low signal levels, by eliminating the effects 
of glitches in peak detector and/or integrator circuits.

-Howard  