Several factors can change the relationship between the indo-1 ratio and
intracellular ionized calcium concentration, [Ca(2+)]i. If you suspect that
this relationship is being affected by experimental conditions, then you
should perform your (indo-1 ratio to [Ca(2+)]i) calibration more than once
(see Carl June and Peter Rabinovich for methods to perform this
calibration). In the end, all of your data is expressed as [Ca(2+)]i in
units like nM, and then meaningful comparisons can be made.

Good luck,
Eric

>Greetings,
>
>We are running Indo-1 for calcium flux plus PE and PE-Cy5 for surface
>markers on an EPICS 753.  When samples are run that have been incubated
>with a drug of interest, we get a higher baseline reading for the indo-1
>ratio.  If we can prove that the drug is causing the higher baseline
>through increased background fluorescence in one of the indo-1 channels, is
>it proper to normalize the ratio baseline by either adjusting PMT voltages
>or sliding the mean ratio histograms so that the baselines match or none of
>the above?  Thanks.
>
>
>------------------------------------------------
>| Justin Fishbaugh                             |
>| University of Iowa                           |
>| Flow Cytometry Facility                      |
>| 48 EMRB                                      |
>| Iowa City, IA  52242                         |
>|                                              |
>| justin-fishbaugh@uiowa.edu                   |
>| http://www.medicine.uiowa.edu/flowcytometry/ |
>------------------------------------------------


/\/\/\_ Eric Van Buren, aa9080@wayne.edu
\ \ \   Karmanos Cancer Institute and Immunology & Microbiology
 \_^_/  Wayne State University, Detroit, Michigan, USA



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:54:12 EST