Re: Scatter of human PBMCs on Vantage

From: Mario Roederer (Roederer@drmr.com)
Date: Fri Oct 08 1999 - 17:01:38 EST


Tina--note that proper compensation does not require the whole 
negative population to be on-scale:  it only requires that the MEDIAN 
of the negative population be on-scale.  See 
<http://www.drmr.com/compensation>.

If you are not used to running PBMC, you may be surprised by the 
relatively low autofluorescence that they have, making it indeed 
impossible, at times, to push all cells of the axis.  Just set the 
voltage so that the top of the negative population is near the top of 
the first decade--you should be OK then.

mr