RE: Acute leukemia

From: Chapple, Peter (peterc@petermac.unimelb.edu.au)
Date: Fri Sep 10 1999 - 02:54:37 EST


Paul,
 
We use 3 colors routinely for leukemia work with the CD45 v LSS gating
technique. CD45 (PerCP or PE-CY5) is
added to EVERY tube - leaving you the Fl1 and Fl2 detectors to add FITC and
PE conjugated antibodies (so in 
effect we are doing 2 color typing and using the third as a gating
antibody).
 
You should read Michael Borowtiz's paper on CD45 v LSS for the background
(Am J Clin Pathol 1993;100:534-540)
and also get a hold of Greg Stelzer's paper in Annals of the New York
Academy of Sciences (March 20 1993) - 
since this has the best "map" of where all the cell types fall (pp.268).
 
Panels are always a contentious issue - I like the paper by the BCSH task
force - J Clin Pathol 1994;47:777-781,
but it's a little out of date now - remember to include the fashionable
marker CD117 (see Bene etal Blood Vol 92 No 2
1998:596-599). However I don't like setting up a screening panel followed by
a secondary panel since it involves
double handling - and labour is the most expensive component of everything
we do. So we have created "one shot"
panels based on the published sources, with an emphasis on the patient
population we see (mostly adult AML).
 
Hope this helps.
 
Peter Chapple
Melbourne AUSTRALIA

-----Original Message-----
From: Paul Kuon [mailto:pmk3351@usl.edu]
Sent: Thursday, September 09, 1999 9:17 PM
To: cyto-inbox
Subject: Acute leukemia


How does one pursue an acute leukemia using three color analysis? How do you
know where to look for the blast population, if the blasts are < 10% of the
total? Do you use a
CD45 Vs LSS histogram setup?  We are new at this particular aspect and would
appreciate any help available.  Also any specific panels that are run for
ALL ,AML,AMML etc. Thanks in advance.
 
Reply to: pmk3351@usl.edu <mailto:pmk3351@usl.edu>  or the group.



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:53:58 EST