Re: FACS Vantage - FL2 Filter Choice for PE

From: FACS_COPY@wehi.edu.au
Date: Fri Aug 27 1999 - 02:54:56 EST


Re Lora Barsky's advice:

> I kick myself for not checking the filter sets earlier.  I discussed these
> issues with my field engineers and we agreed that a 575BP26 should work for
> PE, but really, it doesn't.  I was loosing a lot of sensitivity!!!  
> 
> Hoping others may learn from my experiences, I'm giving the following
> advice:  be sure to run PE with 585BP42 filter set in FL2 on the
> FACSVantage, when using an Enterprise Argon/UV laser.

We use a 575BP26 filter for FL2 on our FACStar because of the presence
of a dye laser (as described by Marty Bigos) but recently went testing
our PE sensitivity.  Swapping in a 585BP42 filter did indeed give us
more signal but also proportionately more background so that the
signal-to-background was _not improved_.  I guess we could have
expected this since the narrower filter is on the PE emission peak
where the signal-to-background is highest.  The wider filter gets more
of the "skirt" region where the signal-to-background is lower. 

Frank.