From: Barsky, Lora (LBarsky@chla.usc.edu)
Date: Mon Aug 23 1999 - 17:02:37 EST


Dear FACS Vantage Operators/Users,

We have discovered recently that we had been running our Vantage with a
different filter set on FL2 than what is issued on both the FACScan and the
FACSCalibur.  We had been running with a 575BP26 for quite a while to
analyze PE.  All of our work "seemed" OK; bright markers were only slightly
dimmer than what was seen on the analysis machines.  But when we needed to
do some dim marker sorts and were coming up with negative expression, we had
a problem. So, we ran some tests by carefully placing the isotype control in
the same channels on all machines.  The Vantage (with the 575BP26 installed)
gave us results that were dimmer than the Calibur by 192 channels! Then when
I installed a 585BP42, which is the same filter used in the Calibur, I was
only 62 channels dimmer.  The 62 channel difference was also seen on FL1, so
I felt the drop was due to laser alignment and stream in air issues.  What a
relief!!

I kick myself for not checking the filter sets earlier.  I discussed these
issues with my field engineers and we agreed that a 575BP26 should work for
PE, but really, it doesn't.  I was loosing a lot of sensitivity!!!  

Hoping others may learn from my experiences, I'm giving the following
advice:  be sure to run PE with 585BP42 filter set in FL2 on the
FACSVantage, when using an Enterprise Argon/UV laser.

Thanks for your time,

Lora Barsky
Children's Hospital Los Angeles
Research Immunology/BMT
323 669-5935