Forwar Scatter is often not even Forward Scatter was RE: Very basic question

From: Robert C. Leif, Ph.D. (rleif@rleif.com)
Date: Wed Aug 04 1999 - 20:21:02 EST


From: Bob Leif
To: cyto-inbox

Firstly, the compromise of employing a laser to both excite fluorescence and
produce a very narrow beam for forward scatter is almost always skewed to
enhance the measurement of fluorescence. Forward scatter really requires
measurements within the forward lobe of the light scatter. This requires a
laser beam with a very small divergence.

Electronic cell volume is still the best way to measure cell volume. For
non-spherical cells, The difference in vertical and horizontal light
scattering can then be employed to determine the shape factor.

"Apparatus and Method for Determination of Individual Red Blood Cell Shape"
assigned to BeckmanCoulter, R. S. Frank, J. L. Wyatt, W. Gong, C. M.
Rodriguez, and R. C. Leif. 5,798,827 (1998).



-----Original Message-----
From: David Lloyd [mailto:lloyddr@novell2.bham.ac.uk]
Sent: Tuesday, August 03, 1999 5:58 AM
To: cyto-inbox
Subject: Very basic question



Dear all

Please excuse this very basic question, I would be very grateful for any
wisdom you can give.

It is generally accepted that "forward scatter = size" within certain
limits;
shape, refractive index etc.

Does this knowledge come solely from the theoretical physics/optical
properties
of particles in laser beams, or is there any empirical evidence correlating
fs
with coulter volume (or any other size measurement)? I have looked for such
evidence but have not been able to find any; perhaps it is further back in
history than I can search.

If anyone knows of any published empirical evidence I would be very grateful
if
you could share it with me.

Thanks in advance



David R Lloyd
University of Birmingham
England

t; +44 (0)121 414 5264/7
e; D.R.Lloyd@bham.ac.uk
http://www.bham.ac.uk/chemeng/actg/actghome