Re: dumb questions

From: Richard McFarland PhD, MD (mcfarland.richard@pathology.swmed.edu)
Date: Sat Jul 24 1999 - 17:22:14 EST


Flowers-

I agree with all of Maryalice's astute comments and would like to add one
comment of my own to the discussion.

When we are evaluating B-cell malignancies we routinely run the following
two four color tubes (in addition to others not directly pertinent to this
discussion)

mK/mL/CD5/CD19;  and
pL/pK/CD20/CD38.

We find as Maryalice suggests that the different  sets of light reagents
can allow you to make a diagnosis when one set of reagents might not. We
use one monoclonal set and one polyclonal set and we switch the
fluorochrome- light chain combinations between the two tubes becasue we
find that the PE reagents tend to pull the light chain positve cells away
from the 45 degree line of non-specificity slightly better than the FITC
reagents.  Also as Maryalice points out CLL's have characteristically dim
surface immunoglobulin staining. The other common small, mature B lineage
CD5+ malignancy is mantle cell which characteristically has bright
immunoglobulin staining. This distinction is not adequate alone to make the
diffirential diagnosis, but it is helpful.


Richard McFarland
Instructor
Division of Hematopathology and Immunology
Department of Pathology
UT Southwestern Medical Center at Dallas
Dallas, Texas