Re: centrifugation

From: Paula Lavery (plaver@po-box.mcgill.ca)
Date: Thu May 27 1999 - 08:56:48 EST


We have been using Fisher microcentrifuges routinely for years. We spin 1 ml
tubes at approximately 5000-6000 rpm (2400-3400g) for 1 minute (setting # 4
or 4.5). We have never had any problems at these speeds and have processed
many different types of cells at these conditions.

_____________________________________________________
Paula Lavery
Transplant Immunology Laboratory
McGill University Health Centre
Royal Victoria Hospital
Montreal, Quebec, Canada
plaver@po-box.mcgill.ca

"If we knew what we were doing it wouldn't be called research, would it?"
-- Albert Einstein


-----Original Message-----
From: Nancy Gin <nancy_gin@yahoo.com>
To: cyto-inbox
Date: Friday, May 21, 1999 1:32 PM
Subject: centrifugation


>
>Hello, We're wondering what everyone out there in the flow world is
>using as centrifuges for immunophenotyping and at what speed. We were
>asked to use the serofuge for centrifugation to save time but we're
>concerned about excess speed that may cause clumping and other
>artifacts. We're currently centrifuging our samples at 340g - 1250rpm
>for 5 mins.with brake on at "2". The CDC MMWR recommends no more than
>400g. The Lo setting on our serofuge is at 500g - 2400rpm. We would
>appreciate very much for any input. Thanks in advance. Nancy Gin -
>United Hospital, St. Paul, MN
>
>_____________________________________________________________
>Do You Yahoo!?
>Free instant messaging and more at http://messenger.yahoo.com
>