An anecdotal report with APC/Cy7 detection is that a 710 SP DC with a 780BP40 did not give any signal when APC/Cy7 was excited by 35mW of 633nm light, whereas it does work well with a 700LPDC and 780BP40 when excited by 647nm light. One recommended filter combination (though not tested by me) is 710 LPDC and a 750LP filter in front of the PMT. You may also want a PMT with enhanced red-sensitivity. A Hamamatsu R3896 has better performance at the red end than the commonly used R928. Dave Coder dcoder@u.washington.edu ----- Original Message ----- From: Joseph Trotter <trotter@scripps.edu> To: cyto-inbox Sent: Wednesday, April 28, 1999 6:35 PM Subject: Re: Dead time > > > Dear Sasha, > > I forgot an important point - we have 4 PMT detectors on laser 1, and > have measured PE-Cy7 off the 1st laser as well as APC-Cy7 off the > 2nd laser. We use the same 780/60 BP filter for both tandem conjugates > involving Cy7 and the 710 LP dichroic as well. I have not tested a 710 SP > dichroic for Cy7 reagents. You ought to use a BP in addition to your > mirror to minimize leakage from other areas of the spectrum. We have left the > standard 610 SP in place to reflect signal to the 710LP dichroic. > > Sorry for the omission, > > Joe