Re: Pharmingen SAV-Phar Red antibody

From: BIGOS@Beadle.stanford.edu
Date: Fri Apr 30 1999 - 07:59:59 EST


Patrica-

I am assuming, from what B-D suggested, that you will 
use the PharRed (Cy7APC is the generic name) with APC.

Then, depending on which PMT position you will be using 
for collecting Cy7APC, you will need either a short pass 
or long pass beam splitter at 710 nm. The bandpass 
filter 780/60 will work well for Cy7APC.

When you specify the beam splitter, you need to mention 
that it will be mounted at a 45 degree angle. You may 
need to purchase a beam splitter holder from BD if you 
don't have an extra or one you can scavange.

We have ordered filters from Chroma Tecnology, also in 
Vermont. Their phone # is 800-824-7662. Similar filters 
that we have ordered have cost from $200 to $300 each.

For best results using Cy7APC it is necessary to have a 
red sensitive PMT. The Hamamatsu R3896 have worked well 
for us. I do not know what B-D supplies in their 
instrument.

Good luck!

-Marty Bigos
Stanford Shared FACS Facility