Dear Ann,

Re 1.
I must admit that I have a somewhat unique optical configuration 
thanks to the guys from BD. I have 3 laser lines conf. with 3-way 
beam-splitter. FL-1 and FL-2 are at the end of the main line, then only 
FL-3 to the right (nothing at the opposite) and then 2 PMTs from UV to 
the left and 2 PMTs to the opposite that I can use from the 633 laser 
(between them is a 3-way beam splitter). I thought, if I use DM 710 SP 
in the place of the 3-way BS and transmit <710 nm to FL1,2 and 3 and 
reflect >710 to the left to FL-4, or I thought to assign it to FL-6 due 
to possible compensation with FL-3.

Re 2.
First, I think it is important to detect RBCs if you don't want them 
in collection tubes, because if the instrument does not detect them 
they can end up in the waste as well as in the collection tubes (am I 
right?)
Secondly, I must go to the log FSC to register all RBCs and then 
(especially on high pressures) I have to play with FSC and even SSC 
obscuration bars to eliminate, as you said, monstrous noise from the
interfering (then 65 kHZ) drop drive.
It could be that because of wider obscuration bars for high-speed 
sorting, FSC has been diminished on my instrument.
Unfortunately, researchers have broken collection tubes in a fuge so I 
don't have any information on cell numbers, but next time I am going 
to take a sample of sorted cells before I give it to them.

Kind regards,
Sasha.


On Tue, 27 Apr 1999 14:18:44 +0000 Ann Atzberger 
<Ann.Atzberger@EMBL-Heidelberg.de> wrote:

> Hello Sasa,
> 
> I also have a Vantage with an optical configuration for 2 lasers. The Fl-3
> and Fl-4 PMT's are at opposite ends to each other. So with the help of a DM
> I can only reflect the emitted light of one laser to only one of these
> PMT,s. Am I right flow-ers??
>  Are you using two 488 lines? Is the optical configuration on your machine
> different?
> 
> On your 2nd question:
> 
> using linear settings on FSC/SSC should give you a very nice (typical)
> profile distinguishing granulocytes, lymphs, rbc,s. Threshold about 60-100.
> When I put FSC on log with ND out, I get a monstrous noise signal. I avoid
> working with FSC log as half the time I can't distinguish between the noise
> and the signal from the cells.
> The abort rate depends also on the size of the population to be sorted. If
> there's a huge amount of RBC's you could try pushing up the threshold level
> above the signal
> of the RBC's. You do not need to be able to see the RBC's to sort the
> granulocytes.
> You dfid not mention purity or recovery.
> 
> Is the fl loss on the turbosort only? which fluorochrome is it?
> 
> 
> Regards
> Ann
> EMBL
> Heidelberg

----------------------