What you measure in all cases is not viability but membrane integrity by dye exclusion. This can be impaired temporarily by electro or chemo-poration processes. Also solvents or active pinocytosis can interfere with dye exclusion properties. In the later case dye retention might be an alternative way of measurement. Another point to consider is the time and dye concentration used. Both, long staining time or high concentration will eventually lead to staining. As a more aggressive scheme you might want to add some DNAse. If that can nibble you DNA of the cells away you know that the cells definitively had it. Regards Gerhard ______________________________ Reply Separator _________________________________ Subject: Viability / nuclear stain Author: suedemag@uci.edu at INTERNET Date: 09/03/1999 20:29 Has anyone used anything other than PI to do viability staining on the FLOW? 7-AAD? I have a researcher using PI and AO and they co-stain ALL the cells, virtually 100%, in what they believe to be live samples of Chondrocytes. Any hints? Does the collagenase used to break up the matrix of the cartilage cause permeabilization? I'm kind of at a loss to explain to him what is happening. Any better dye for this application? Thanks in advance! Susan DeMaggio, MS BSMT(ASCP)QCym Resource Manager Optical Biology Shared Resource University of California, Irvine Irvine, CA 92697-2275 phone 949-824-4110 fax 949-824-3571 <http://mamba.bio.uci.edu/~suedemag/OBCore/index.htm>http://mamba.bio.uci.e du/~suedemag/OBCore/index.htm