Re: (Fwd) 'alternative' live/dead stains?

From: Mario Roederer (Roederer@Beadle.stanford.edu)
Date: Wed Jan 13 1999 - 12:33:30 EST


You should have no problems at all using PI in a 3-color system; however, your
other two colors should be FITC and PE.  PI's emission is strongest in the FL3
channel; you can then compensate between FL3 and PE (FL2) to remove the
contribution of PI into the PE channel.  Note you will probably have to turn
down the voltage on FL3 significantly, because PI is so bright (or, you could
titer the PI to lower concentrations, but I wouldn't recommend this).

Note that we frequently do 3-color + PI experiments--i.e., stain with FITC, PE,
and Cy5PE, and PI as well.  The PI (in dead cells) is so much brighter than
nearly all Cy5PE stains that the dead cells can be easily excluded (and you
don't have to worry about PI compensation, since you are excluding the PI+
cells).

However, since you wish to enumerate PI+ cells, you need to devote one channel
to its measurement--hence, use the FITC/PE/PI setup noted above.

By the way, I recommend against 7-AAD; it's not as bright as PI.  However, you
might want to compare using it also in a FITC/PE/7-AAD set up.