   I am not an expert in the use of PAC-1, but I think there has been
some misinformation on the list about it and I would like to put in my 2
cents.

PAC-1 is an IgM monoclonal that binds to an epitope of GPIIb/IIIa which
appears upon activation.  The appearance of the epitope requires that
GPIIb/IIIa be in its normal configuration and thus requires the presence
of calcium.  One cannot successfully use PAC-1 on platelets that have
been treated with EDTA (citrate is OK), or with fixed platelets
(fixation seems to induce artifactual binding of PAC-1 to both resting
and activated platelets).

The epitope recognized by PAC-1 is not the fibrinogen binding site.  As
far as I know, there should be no competition between PAC-1 and
fibrinogen.  Other monoclonals, such as 7E3 do recognize the fibrinogen
binding site and will compete with fibrinogen, but I don't think that is
true of PAC-1.

I hope this is helpful.

--------------------------------------------------------------------------