Cleaning + background events

From: Derek Davies (daviesd2@icrf.icnet.uk)
Date: Tue Dec 02 1997 - 02:59:35 EST


Hello fellow fans of flow!

I apologise in advance but I would like to return to the ever-popular
cytometer cleaning question.

I have a FACScan and a FACS Calibur, both multi-user high throughput
machines, both used for DNA and for phenotyping experiments. I have
impressed upon all my users that after running their samples they MUST run
a cleaning procedure which is:

1. dH2O 5 mins on high flow
2. detergent (7X) 5 mins high flow
3. dH2O 2 mins high flow (if PI has been used I put a 5 min bleach step
subsequent to this)
4. Standby (If I am about I tend to leave it on run)

I also run detergent and dH2O through the entire fluidics system once a
week. So far so good, rarely get clogs, bead and DNA CVs and profiles
fine, haven't had to slay too many users for forgetting and I am happy
(well as happy as I get these days). Recently though I have been informed
of a few 'problems'. I have one user in particular who quite often ends up
on the linoleum in a black apoplexy. She is running 3 or 4 colour fetal
thymocytes. Before she starts her run she acquires 'data' from a tube of
autoclaved distilled water. Her complaint is that events are acquired that
fall within her cell gate. So.....my question is this:

Should I expect to see 'cells' in a tube of dH2O? If so, how many, say,
per minute? I have just run such a tube on high flow rate and got over a
10 minute period an average of 2-4 events per second (Low and high flow)
in a typical cell gate. I would be tempted to say that this is
insignificant in a real sample compared with the cell flow rate.

Does anyone have a view on what is acceptable (I checked the archives as
there was a flurry of cleaning correspondence earlier in the year but
couldnt find the answer!)? Am I wrong in saying that there would never be
a *totally* clean background?

Oh I use PBS as sheath fluid and I have changed the in-line filter.

Thanks in advance!

Derek


****************************************************************************
*  Derek Davies                       Voice: (44) 0171 269 3394            *
*  FACS Laboratory,                   FAX: (44) 0171 269 3100              *
*  Imperial Cancer Research Fund,     e_mail: derek.davies@icrf.icnet.uk   *
*  London, UK                                                              *
*                                                                          *
*  Web Page: http://www.icnet.uk/axp/facs/davies/index.html                *
****************************************************************************



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:50:23 EST