Re: Lost cells

From: Ray Hicks (rh208@cus.cam.ac.uk)
Date: Thu Oct 23 1997 - 06:41:24 EST


>
>Where are 80% of the cells going every time?  Is there some physical
>/ fluidic / optical explanation for why they may not be counted on
>the FC?  Or, is there a greater chance that they may be lost prior to
>acquisition; if so, how?
>
>Please respond with even remotely possible explanations, especially
>personal experience.
>
>

Hi Darren,

You might find that they get blown into the void volume of the peristaltic
tubing when they're first put on, try switching to standby when the tube
appears dry, this drops the pressure in the tube and allows the peristaltic
tubing to relax and eject the cells that were forced in, they'll get
diluted a little.  Alternatively try removing the outer metal tube from the
sample port (loosen the knurled nut a little and slide the tube off) this
would allow you to rule in/out whether the drop uptake was significant.

Ray