Culture 1-2 x 106 cells/mL in 1 mL CGM
(RPMI + 2% each pen.-strep. and AB serum) with desired stimuli/toxins
for 24 to 48 hours.
-
Spin cells down and dump supernatant.
-
Wash* once.
-
Fix 20 minutes, room temperature, with 1 mL 4% paraformaldehyde
in PBS.
-
Spin & dump.
-
Permeablize 10 minutes with 0.15% Triton X-100/0.1%
BSA in TBS.
-
(Seki uses 0.1% Triton X-100/0.1% BSA in TBS for 5 minutes.)
-
Spin & dump.
-
Wash once.
-
(Seki washes a total of three times at this point.)
-
Stain in a total volume of 100 mL for 30 minutes (dark/4C).
-
(Seki stains for 20 minutes. Other protocols call for including
detergent in the staining mixture.)
-
Raise volume to 2 mL and wash.
-
Dump supernatant, raise volume to 500 mL, and read the same day by flow
cytometry.
-
For FITC conjugated antibodies, excite with 488 nm and read fluorescence
at 530 nm.
* Washes: 1800 RPM, 8 minutes, 4C, in 2 mL PBS.
Controls should include unstained/"cells only", FITC-IgG1/"conjugate
control", and FITC-bcl-2 labeled unstimulated/untreated cells.
For both Dako FITC-bcl-2 (cat #F 7053, 100 mg/mL, F:P=4-5) and BDIS
FITC-IgG1 (cat #349041, 50 mg/mL, F:P=?),
use roughly 2 mL of Ab solution per tube. (Dako calls for 10 mL;
BDIS calls for 20 mL.)
Based on Seki, H. et al. Cellular Immunology 163,
30-36 (1995)
|