There is a letter to the editor in BLOOD , September 1996,
15:88(6):23558-61 by van Vugt, MJ et al which shows that PE-CY5
binds SPECIFICALLY to the Fc(gamma)R1 receptor (CD64). These
experiments using mouse B cells transfected with different human Fc
receptors showed that only FcR1 cells bound PE-CY5. Whole blood
showed much less binding than isolated cells. Their recommendation
was that any antibody that you are using labelled with PE-CY5 be tested
for FcR1 binding.
In our hands blocking with heat inactivated normal human AB serum will
significantly reduce the Fc binding in humans.
Mike Keeney