Re: PI Staining

Ryan Hung (rhung@vcn.bc.ca)
Thu, 10 Jul 1997 15:18:28 -0700 (PDT)

On Wed, 9 Jul 1997, John Ladasky wrote:
> Unlike monoclonal antibodies, PI has a fairly high off rate. So
> having excess stain in the solution helps to keep all of the binding sites
> of the DNA saturated, thus facilitating DNA quantitation. If you just want
> to distinguish live cells from dead cells, you can probably get away with
> washing away the excess PI. But not for long -- after a few hours the dead
> cells would not fluoresce significantly more than the live ones.

Hmm, is this true for unwashed cells as well? I've been using a one-step
hypotonic lysing buffer that contains PI (50 ug/ml), but I've been varying
the time (up to several hours) before reading on the FACS scanner. As
such, would the PI saturate all nuclei, giving a reduced amount of
hypodiploid fluoresence and sending everything up to the G0/G1 peak (thus
giving abnormally low levels of apoptosis)?

Ryan.

_/ \__/ \__/ \__/ \__/ \__/ \__/ \__/rhung@vcn.bc.ca__/ \__/ \__/
\_Apoptosis=programmed cell death/ \__/ \rwhung@netinfo.ubc.ca \__/ \__
_/ --you can't live without it!/ \__/ \http://www.vcn.bc.ca/people/rhung
\__/ \__/ \__/ \__/ \__/ \__/ \__/ \My words Copyright (C) 1997 \__


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu