RE: ANCA by flow

Edoardo Colombo (coledoco@tin.it)
Mon, 14 Jul 1997 07:27:13 +0200

John
the best way I know to recover neutrophils is to lay
0.5 ml peripheral blood onto a 1 ml ficoll 1.077 and
let motionless for 15 minutes : you will have a clear
pellet containing all WBC and the platelets (by the
way this is a quick and simple way to obtain very rich
plt specimens); this tecnique do not activate the PMN
at all.
Regarding the permeabilization I think that Triton X-100
(working concentration = 0.1 %) is the best item you
can use
Finally I think that Fab FITC IgG , as used by me to
detect anti-platelets autoantibosies , is the right way
to detect the target you need.
Best regards

Edoardo Colombo MD
Director Flow Lab
Dept. of Pathology
Valduce Hospital
COMO , Italy
coledoco@tin.it


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu