>our question is whether using ionomycin (as our positive control/at various
> concentrations) should the ratio of violet to blue return to near baseline
> after the influx of calcium?
>
In our hands, working mainly with mouse thymocytes, the ratio stays high
after the addition of ionomycin until such a time as the cells begin to
become "less healthy" as indicated by a decreasing forward scatter, slightly
increased side scatter. If you let the time course continue you can plot
nicely the return to near baseline of indo-1 fluorescence ratio correlating
with scatter change in a 3D type plot.
As a rough method of duplicating this effect, try adding a squirt (technical
measurement I use) of 70% ethanol to the end of one of your sample time
courses and you should see a similar pattern to the above ionomycin
treatment. This doesn't hang around in the sample lines, and can be a quick
way of checking whether you've got any loading if ever you don't get a
response from your agonist.
Hope this helps,
Geoff
======================================================================
Geoffrey Osborne | ____ __ o Ahh!
Flow Cytometry (FACS LAB) | __ `\ <,_
John Curtin School of Medical Research, | __ (*)/ (*)
Australian National University, | ==============|
CANBERRA, AUSTRALIA. | |--|
Email: Geoff.Osborne@anu.edu.au | |--|...
Phone: 61 6 249 3694
FAX: 61 6 249 2595
-----Surfing the Web?: Try http://jcsmr.anu.edu.au/facshome.html------
======================================================================
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web