Old Parformaldehyde

Calman Prussin (Calman_Prussin@d10.niaid.pc.niaid.nih.gov)
Mon, 06 Nov 95 15:27:15 EST

I think it depends on what you are using the paraformaldehyde (PFA) for. The
constraints on using PFA for fixation of already stained cells (for later
analysis) is less than that for uses with intracellular staining. For the
former, I think PFA in the frig for months should be fine.

A lot of the fears may simply be that, however a number of investigators I have
talked to make up fresh PFA for each use. I think the concern is old PFA causes
overfixation leading to increased nonspecific binding.

Considering how much trouble it is to make fresh PFA, we make up a couple of
litres at a time, aliquot them into 15 ml tubes and freeze the whole lot. I
will use thawed PFA for a week or two after thawing. I have heard that Ulf
Andersson's group in Stockholm use a similar approach, except they don't hold
on to the thawed PFA for 2 weeks.

Calman Prussin
Laboratory of Allergic Diseases
NIAID, NIH

---------------------- Replied Message Body ----------------------
Date: 11-2-95 6:48am
From: {herro001@maroon.tc.umn.edu}:unix:niaid
To: calman prussin:10:niaid,
karin hartmann:10:niaid,
barbara butcher:4:niaid,
ramya gopinath:4:niaid
Subj: Old Parformaldehyde
Also-to: cytometry mailing list <cytometry@flowcyt.cyto.purdue.edu>

--------------------------------------------------------------------

I have read the warnings to only use "fresh" parformaldehyde for fixation, but
I
don't recall ever hearing why.

Could someone explain the sorts of artifacts that old paraformaldehyde will
introduce? And what exactly IS old?


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu