Ira Schieren (75354.1242@compuserve.com)
12 Oct 95 14:22:18 EDT

I successfuly detect 90 degree light scatter using a 370nm long pass dichroic
filter under the detector of my Epics 753. I usually run my argon laser at 75mw
for UV. The light scatter signal is not as strong as it with the laser at 488,
but it is quite usable. All you have to do is double or tripple the voltage on
that detector. Some people have used a microscope cover glass in place of the
dichroic filter with success. Good Luck!

Ira Schieren
Howard Hughes Medical Institute
Columbia University
Cell Sorter Facility
212-305-2704
75354.1242@CompuServe.com


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu