Re: Right angle scatter with UV laser line?

Joseph Webster (J.Webster@centenary.usyd.edu.AU)
Thu, 12 Oct 1995 08:53:15 +1100

At 11:06 AM 11/10/95 -0500, Raymond B. Hester wrote:
>
>We don't do that much UV on our flow but I've just across what seems to
>be a problem and I knew someone out there could help.
>
>Do you not get right angle light scatter (or very little of it) when
>using the UV line? I was just setting up for an experiment in which we
>will be using UV on our FACS 440. For alignment, I was using some old
>Polyscience beads
>(about 5 micron dia.) that still fluoresce and I didn't see much side
>scatter.

Two things spring to mind:
When I did some UV work on a FACS 440 some years ago, I found the side
scatter to be MUCH lower sensitivity than with visible laser; I think that
UV light is not scattered as much, and the PMT was less sensitive to UV.

I did not get ANY UV side scatter until I removed the 488 bandpass filter
from the side scatter detector! (took me at least a week to think of it...)

Good Luck, Joseph.

Joseph Webster (O.I.C. Flow Cytometry & Communications) ===
Centenary Institute of cancer Medicine & Cell Biology / \
Locked Bag No.6 || o o ||
Newtown, NSW 2042 || | ||
AUSTRALIA. | \___/ |
Ph: 61-2-565-6110 \ /
Fax: 61-2-565-6101 |||||
E-Mail: J.Webster@centenary.usyd.edu.au


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu