Subject: Re: PE/Cy5 on a FACS Calibur
Author: steven_micko@email.eushc.org at @UCSD
Date: 9/25/96 11:03 AM
Calman,
We are a beta site for the second (red) laser on a FACSCalibur. I do my
daily QC with Calibrites (FTC, PE, blanks, PerCP and APC), and it works
fine. Our biological work uses PerCP, APC, and Cychrome, and I have had
no complaints. I don't know about PE/Cy5; is it the same as CyChrome?
The present status of 4 color work on the Calibur is that it is in Beta
testing. We use cross-beam compensationS between FL1 & FL2, FL2 & FL3,
FL3 & FL4.
Steve Micko
Emory Hospital
Atlanta
______________________________ Reply Separator _________________________________
An associate has asked me a third hand question regarding the utility of
using PE/Cy5 conjugates on a FACScaliber in 3 color applications.
Questions for the group and for BD lurkers in particular:
1. Do PE/Cy5 conjugates work any better or worse on a Calibur than on a
FACSCAN in 3 color/ 1 laser applications.
2. What is the present status of 4 color work on the Calibur? Any plans
for cross-beam compensation to allow simultaneous use of FITC, PE,
PE/Cy5 and APC?
Thanks,
Calman Prussin
>-- Saved internet headers (useful for debugging)
>Received: from flowcyt.cyto.purdue.edu by mail.ucsd.edu; id QAA28921 sendmail 8
>Received: by flowcyt.cyto.purdue.edu (940816.SGI.8.6.9/930416.SGI.AUTO) for cyt
>Received: from drake.eushc.org by flowcyt.cyto.purdue.edu via ESMTP (940816.SGI
>Received: from email.eushc.ORG (email.eushc.org [163.246.32.120]) by drake.eush
>Received: from ccMail by email.eushc.ORG id AA843674586 Wed, 25 Sep 96 11:03:06
>Date: Wed, 25 Sep 96 11:03:06 E
>From: "steven micko" <steven_micko@email.eushc.org>
>Encoding: 1153
>Message-Id: <9608258436.AA84367
>To: Cytometry Mailing List <cyt
>Subject: Re: PE/Cy5 on a FACS Calibur
![]() |
![]() |
![]() |
![]() |