Re: laser current

Howard Shapiro (hms@shapirolab.com)
Wed, 19 Jun 1996 23:22:06 -0400 (EDT)

A new laser running in light-regulated mode will draw higher than normal
current when it is out of optical alignment. Yours may have gotten jostled
in transit.

While I am not even sure whose lasers B-D is using in FACScans these days,
many "adjustment-free" low-power air-cooled argon lasers of the type used in
clinical instruments such as the FACScan do have some provision for mirror
adjustment, and I have had to do this on occasion when a laser came from the
factory. "Walking" the mirrors reduced the current draw.

Is this an option?

-Howard


Home Page Table of Contents Sponsors Web Sites
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu EMAIL robinson@flowcyt.cyto.purdue.edu