I don't think your formula is obvious, although it is suprisingly simple and
although I have seen the formula before (in glare correction in microscopy,
where pixels - due to stray light - pick up light when they are dark, but loose
light to neighbouring pixels by the same mechanism as well).
_______________________________________________________________________________
Subject: background subtraction
From: Alice.L.Givan@dartmouth.edu (Alice L. Givan) at INTERNET
Date: 11/7/95 10:13
Here's an old problem: What do people do about subtraction of isotype control
values from flow determination of the so-called "% positive" population? In
most lab with which I am acquainted, if the isotype control marker is set at
3%, then 3% is subtracted from every subsequent sample. Although I have never
heard it discussed in public, I have known for a long time that this approach
is obviously incorrect -- as can be seen if one imagines a sample with clearly
ALL the cells brightly stained. By subtracting the control, it would then
appear that only 97% of the cells were stained. The problem with this
"classical" approach is that we should be subtracting 3% of the true negative
cells only because those are the ones that are mistakenly contributing to the
determination of the positive population.
When I follow this through algebraically, I come up with the following
formula:
y=(x-c)/(1-c/100)
where y is the actual % positive
x is the measured % positive
c is the % positive set for the isotype control
Am I wrong about this (or is it so obvious that everyone else does it already)?
Does someone have the patience to check my derivation?
Thanks.