Hello,
As our only flow cytometer is usually busy acquiring data, I installed
the WinMDI software and paraphenalia on a PC. Though luck, though,
becuase I can't get the HPdir and HPcopy prg's to copy the data files
from drive to harddisc. It keeps returning "error 158". Hints, anybody?
Tech specs: B-D FACScan, HP series 300 type 9000 computer with HP9153C
disc drive, running under HP pascal 3.0. The PC is a standard 486 DX-2
(66 MHz) clone, with standard highdensity 3.5" drive.
Thanks,
-------------------------------------------------------------------------
Guy Hermans
-------------------------------------------------------------------------
Very usefull replies included:
>From pif@box-p.nih.govMon Apr 22 11:07:27 1996
Date: Thu, 4 Apr 1996 08:28:10 -0500
From: paula fukushima <pif@box-p.nih.gov>
To: cyto-inbox
Subject: Re: HPCOPY and HPDIR problems
>
>I had a similar problem with Oswego and Multicycle. It turned out that a
recently installed LAN was the problem. It was taking up too much HIMEM. I
now have to disinstall the LAN before I can use those programs.
>
>From trotter@scripps.eduMon Apr 22 11:07:34 1996
Date: Thu, 4 Apr 96 06:42:08 PST
From: Joseph Trotter <trotter@scripps.edu>
To: cyto-inbox
Subject: Re: HPCOPY and HPDIR problems
Guy,
HPCOPY and HPDIR will only read HP diskettes formatted by HP MEDIAINIT
program using option 16. Diskettes formatted by LYSYS or Master Page
are not suitable because MEDIAINIT is called with other option numbers.
See http://facs.scripps.edu/hp2pc.txt
Regards,
Joe
>From info@appcysys.demon.co.ukMon Apr 22 11:08:11 1996
Date: Fri, 5 Apr 1996 12:12:30 GMT
From: info@appcysys.demon.co.uk
To: cyto-inbox
Subject: Re: HPCOPY and HPDIR problems
You have found the right team!
Applied Cytometry Systems sell allsorts of goodies
that will allow you to move data from HP to PC or
to collect data directly into your PC and control
your cytometer with a PC.
Checkout our WEB page below for more info and send
us a message.
Please note we will have a skeleton staff from the
10 to 18 the April as All two of us are off to ISAC
and our Admin person is off being a first-time
grandmother!
Terry Godley Tel: +44 (0)1909 566982
Applied Cytometry Systems Fax: +44 (0)1909 561463
Dinnington Business Centre Email: terry@appcysys.demon.co.uk
Outgang Lane, Dinnington WEB: http://www.enterprise.net/appcysys
Sheffield S31 7QY UK
>From hspengl@gwdg.deMon Apr 22 11:09:08 1996
Date: Tue, 16 Apr 1996 23:44:30 +0000
From: Hans-Peter Spengler <hspengl@gwdg.de>
To: cyto-inbox
Subject: HP to PC
Guy,
there are some other useful programs available to copy files from HP to PC.
1. We use CopyHP instead of HPCopy. I could download this program from
"http://amcradio07.amc.uva.nl/flow/flow.htm" but now downloading does
not work. (AMC.UVA.NL is down; site is being reconstructed...GH)
Copy HP 1.0 (ZIP-Archive) (Can only be obtained locally).
Format HP Floppy's and copy HP-Files on a PC.
Requires Microsoft Windows 3.1 or better.
2. In USA (Verity Software or National Instruments) as well as in
Germany (JTES) you can buy a soft-/hardware kit (approx. 1000)
which makes it much faster to copy files from the HP to PC. JTES will
send you a testtrial kit for up to 10 days. It is very easy to
install the HPIB board into yur PC and to run the software (we have
seen this in a lab at University of Regensburg/Germany). In our lab we failed
to connect the HP with the PC (????). A 486/DX33 is very good but the
HPIB cable is limited to about 4 m distance from the HP to the PC
Hans-Peter Spengler
Dept. of Immunology FAX: +49-551-395843
University of Goettingen Phone: +49-551-395819
Kreuzbergring 57
D-37075 Goettingen/Germany
>From hspengl@gwdg.deMon Apr 22 11:08:57 1996
Date: Thu, 18 Apr 1996 09:11:53 +0000
From: Hans-Peter Spengler <hspengl@gwdg.de>
To: cyto-inbox
Subject: Re: HP to PC
Guy,
the address is:
JTES BioTec
Dr. Juergen Treumer
An der Eiche 21
D-24991 Freienwill
Phone: 0461-180542 (private: 04602-1390)
FAX: 0461-180557
There are two different versions available.
1. Just file transfer HP->PC and PC->HP (DM 1.190 plus 15% TAX)
2. File transfer, printer buffer for the HP, and HP printer connected
with the PC (this you needs if you have a HP colour printer at your
HP). Price DM 1.640 plus 15% TAX).
For the PC you can get only a DOS program but it runs under windows
3.1x. You also have to install a file on the HP in Vol. #11. An
english version of the PC and HP software is already included.
I think it is a wonderful solution for transmitting files from the
old 300/310 HPs and it is very fast.
Yours
Hans-Peter
software
Dept. of Immunology FAX: +49-551-395843
University of Goettingen Phone: +49-551-395819
Kreuzbergring 57
D-37075 Goettingen/Germany
----------------------------------------------------------------------
Many thanks to all these people
-------------------------------------------------------------------------
Guy Hermans
PhD Student
Immunology and MS research unit Fysiology Dept., Immun research
Dr. L. Willems-Institute Limburg University Centre
University Campus University Campus
B-3590 Diepenbeek B-3590 Diepenbeek
Belgium Belgium
+32(0)11.26.92.07 Voice E-mail: ghermans@luc.ac.be
+32(0)11.26.92.09 Fax
-------------------------------------------------------------------------
![]() |
![]() |
![]() |
![]() |