Should be easy, I did it lots several years ago!
But...
> The file utilities in LYSYSII don't see the files on the floppy.
This probably means there is a blank filename at the top of the
list on the floppy; If the BD File Utility finds a blank name, it
assumes that is the end of the list, and doesn't bother to look
any further.
> If I go to the HP-Pascal level on Consort32 system, I can see
> the files by listing the floppy dir..
To do this, you must be able to use the Pascal Filer?
If so, armed with a list of files and a workable directory on
the hard disk, you can copy files like this:
c #3:DATA.001,#11:/TOM/DATA/FRIDAY/DATA.001
NOTE the comma! it separates the source from the destination!
Then there are some shortcuts!
You can use "$" to give the copy the same name as original, without
typing the whole thing; e.g.
c #3:DATA.001,#11:/TOM/DATA/FRIDAY/$
You can use wildcard characters to match groups of characters;
e.g "c #3:DATA.=,#11:/TOM/DATA/FRIDAY/$" will match any
files on floppy which start with "DATA." and make copies with
the same names in the specified directory on the hard disk.
> I'm assuming at this point, that if I get them onto the
> consort32 that LYSYSII will recognize them. Am I correct?
Yes, I've done exactly that many times.