Bead CV's
Howard Shapiro (HPANDA@HARVARDA.HARVARD.EDU)
Mon, 06 Jun 94 20:35:46 EST
Bead CV's vary considerably with manufacturer, manufacturing process, and
lot number, among other things. Over the years, the beads I have found to
consistently have the lowest CV's have been 2 micron green spheres from
Polysciences, which come out around 1.2 % in scatter and fluorescence in
the best systems on the best days. I've gotten 1.2 % in a Cytopup with
arc lamp illumination at 436 nm, and in laser systems, including one with
a 441 nm, 100 mW HeCd laser. If the lamp or laser noise is 1% RMS, the CV
won't be any lower than that, in the absence of saturation and bleaching
effects. Not all bead lots are that good.
--Howard