Cell cycle analysis

From: kathy schell (kschell@facstaff.wisc.edu)
Date: Tue Dec 10 2002 - 15:14:42 EST


Hi, Kenichi,
A couple of suggestions:
1.  As has been suggested, centrifugation may be the key.  You have
fixed your cells in MEOH and they are much lighter than before
fixation.  I will demonstrate this to students when they are careless
with this step--look in the supernatant and see how many still are
there.  When we sort live cells stained with Hoechst 33342, we have
no trouble recovering the number indicated on the cytometer.
2.  Are your proteins identifiable with monoclonal antibodies that
would be appropriate for flow?  Looking at cell cycle and protein
expression (cyclins in particular and see Chow, et al for measuring
phosphorylated proteins in Clinical Cytometry) is well described in
Current Protocols in Cytometry and would require no sorting.  We have
found that trying MOABs that may not have been used for flow actually
work there.
Good luck,
Kathy
--
--------------------------------------------------------------
Kathy Schell
Supervisor, UWCCC Flow Cytometry Facility
600 Highland Ave.  K4/535
Madison, WI 53792
Voice:  608-263-0313
e-mail:  kschell@facstaff.wisc.edu



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:31 EST