Re: luciferase essay using flow cytometer?

From: David Galbraith (galbraith@arizona.edu)
Date: Fri Dec 06 2002 - 15:47:37 EST


This question seems to come around at regular intervals!  As Dick notes,
the problem is that even transgenic organisms expressing large amounts of
luciferase do not produce sufficient photons per second for a significant
number to be collected by the flow cytometer during the (very short) period
of time that the cell is in the focal point of the collection lens.  This
is compounded by the fact that the light is emitted in all directions, and
the efficiency of collection of light by the flow cytometer is nowhere near
100%.  If you were to slow down the cells, this would work (as of course
does conventional imaging of luciferase in transgenic organisms under
static conditions!).  however, my guess is the cells would have to be very
nearly stationary....

Best wishes

David


At 08:34 AM 12/6/2002 +0000, Simon.Q.Rice@gsk.com wrote:

>I've tried observing cells using luciferase activity in the cytometer,
>without success.
>Luciferin requires ATP and molecular oxygen to drive the reaction, in the
>presence of Mg2+.  The difficulty I had (I think) was getting the
>substrate (luciferin), and ATP, into the cells without damaging
>them.  Even using a caged luciferin, which is cell permeant and activated
>intracellularly, didn't work.  If anybody's got past that stage or wants
>to discuss further off line, I'd be very keen to hear from them.
>
>Simon
>
>Simon QJ Rice
>GlaxoSmithKline R&D
>Harlow, UK
>
>
>"Richard Haugland" <richard.haugland@probes.com>
>
>05-Dec-2002 03:02
>
>
>
>         To:        "Cytometry Mailing List"
>
>         cc:
>         Subject:        Re: luciferase essay using flow cytometer?
>
>
>Interesting question. The chemiluminescence peak of luciferin is at
>about 560 nm, which is a bit shorter wavelength than phycoerythrin
>emssion. Of course, it does not need any excitation to get the emission.
>But then, I don't know if the rate of emission will be fast enough to
>get enough photons during the transit time that you will be able to
>detect the signal. But then S/N should be great because all photons
>should be "real." I don't know what technical difficulties there may be
>doing sorting with the laser turned off but hope it works.
>
>
>
>
>
>Nan Jiang wrote:
>
> >Dear all:
> >
> >I am wondering if anybody ever sorted cells infected with luciferase
> >plasmid. Or is there a particular wavelengh we can use to detect
> >luciferase activity in cells and sort them based on that.
> >
> >Thanks in advance.
> >
> >Nan Jiang
> >
> >Department of Internal Medicine
> >Cardiology Division
> >UT Southwestern Medical center at Dallas
> >6000 Harry Hines Blvd.
> >Dallas, TX  75390-8573
> >(214) 648-1175
> >(214) 648-1181
> >
> >
> >
>
>
>
>


David W. Galbraith
Professor of Plant Sciences
Department of Plant Sciences
University of Arizona
303 Forbes Building
P.O. Box 210036
Tucson Arizona 85721-0036 USA.

Tel: (520) 621-9153
Fax: (520) 621-7186
Email: galbraith@arizona.edu
http://latin.arizona.edu/galbraith



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:31 EST