Re: GFP sorting question

From: Joanne Lannigan (jl7fj@virginia.edu)
Date: Fri Oct 18 2002 - 12:53:14 EST


Hi Andy:
What size nozzle are you using and how big are your cells? Are they clumpy,
is your abort rate high? If your cells (or cell aggregates) are too big for
your nozzle size you can get exactly what you describe. Your nozzle size
should be at least 5X the largest particles in your suspension. Even if your
nozzle size is appropriate for the cell size, if you have lots of cell
aggregates it can create the same type of problem. Try switching to a larger
nozzle and see if that doesn't help. Joe Trotter did a great demonstration
of this effect at the Bowdoin Course where he had a 70 um nozzle and sorted
5um beads and the streams looked great. He then added some 10 um beads and
it still looked pretty good. When he added 15 um beads we started to see
some degradation in the streams and when he added the 25 um beads it looked
just awful with lots of spray between the center stream and the side stream.
It was enough to convince me that nozzle size is important to get a clean
sort! I always ask my clients what size their cells are (and if necessary
verify what they are telling you with beads of different sizes under a
microscope). Also if they have been activated they may in fact be
significantly larger. If their samples are traditionally clumpy and the
usual remedies don't work I generally will go with a larger nozzle size
(helps prevent clogged nozzles as well). Drawbacks are that you have to sort
at lower pressures and you have larger drops (i.e. larger sort volume), not
to mention you are changing tips and realigning more often as well. But I
must say I have seen a considerable improvement in the quality of these
sorts in subscribing to this philosophy. THANKS JOE!
Give it a try-
Regards,

Joanne Lannigan, MS
Director, Flow Cytometry Core Facility
University of Virginia
P.O. Box 800734
Charlottesville, VA 22908-0734
Office: 434-924-0274
Lab: 434-243-2695
Fax: 434-982-1071
email: joannelannigan@virginia.edu


From: Andrew Oberyszyn <oberyszyn.2@osu.edu>
Date: Wed, 16 Oct 2002 16:45:14 -0400
To: cyto-inbox
Subject: GFP sorting question


Hi FLow-er in Flow land,
I've come across an interesting (and confusing) problem which sorting GFP
cells.  I've noticed several times that I get alot of fanning while sorting.
I don't believe it is the Frequency, drive or phase since setting up the
sort gives me beautiful side streams and great separation.  I first notice
this on my Elite and now I have a FACSVantage and 2 of the last 3 GFP sorts
showed this problem. After todays sort, which had terrible fanning, I put on
a sample of unstained fixed PBL's to see if it was instrument related.  Not
surprisingly, that sample gave great side streams.  Obviously the problem is
with the sample.  I've never (never say never!) seen a fanning problem with
surface stained cells and I don't understand what could be causing this to
happen with GFP cells?  Could the transfection method somehow "charge" the
cells which interferes with the charge applied for sorting?  As far as I can
tell the media (which is MEM with Pen, strep, L-Glut, FCS, bicarb.)
shouldn't do anything?

Any help or suggestions would be greatly appreciated!

Thanks in advance.

Andy


^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
  Andy Oberyszyn, M.S.
 Core lab manager
 The Ohio State University
 University Cell Analysis & Sorting Core
 424 Davis Heart & Lung Institute
 473 West 12th Avenue
 Columbus, Ohio 43210
 Tel: 614/292-FLOW(3569)
 Fax: 614/292-7335
 E-Mail: cytometry@osu.edu
  Website: http://heartlung.osu.edu/hlri/corelabs/flowcore.jsp

^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
"What if the Hokey Pokey is really what it's all about?!?"



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:27 EST