TUNEL assay variability, trivial causes

From: DARZYNKIEWICZ ZBIGNIEW (DARZYNK@nymc.edu)
Date: Fri May 31 2002 - 09:25:38 EST


My previous response to the problem of variability of TUNEL assay was
restricted to the genuine differences in frequency of DNA strand breaks.
However, there are often  trivial, technical reasons for irreproducibility.
One relates to the fact that some investigators collect only adherent cells
that are harvested by trypsinization or EDTA, discarding the medium from
flasks. Because most apoptotic cells detach, they are discarded with the
medium. The "efficiency" of discarding medium may vary from flask to flask,
hence the variability in frequency of apoptotic cells.The correct way, of
course, is to pool the cells from medium with the these harvested by
trypsinization.
Apoptotic cells are very fragile and often completely disintegrate during
centrifugations. Hence, their frequency (apoptotic index; AI) may vary
depending on number of cell rinses, centrifugal force and time, etc. Also,
compared to nonapoptotic cells, apoptotic cells have different propensity to
attach to the tubing. The variability in handling the samples (including the
absolute number of total cells per sample) is then reflected by differences
in AI.