Hi Marcia,
one problem of your sort maybe is autofluorescence in your cells.
A second is clumping, that means you have GFP positive and
negative cells sticking together.
We had the same problem and now sort only the smallest part
(foreward versus sidescatter) of a cell population, which have to
be still GFP-postive. Taking only the small cells, you exclude 1)
cell clumps and 2) big cells with high autofluorescence.
We thereafter get much higher efficiencies (more than 90%
positives).
Hope this helps

Andreas



Date sent:		Tue, 30 Apr 2002 15:45:17 -0500
From:			mlh2207 <mlh2207@ksu.edu>
Subject:		sorting GFP macrophages
To: cyto-inbox

>
> Hello all. We have created GFP macrophages using the pIRESneo vector with the
> EGFP gene inserted into the MCS. The cells were then selected using G418.
> Here's the problem...even with selection we have two distinct populations, one
> negative and one positive. We are now trying to sort the cells, but cannot
> eliminate the negative population. The best sort results that we have
> accomplished is 60% positive and 40% negative populations. Does anyone have
> any clues or suggestions on why this is happening and what we can do to
> reduce/eliminate the negative population?
> We had hopes of using these cells for in vivo studies.
> Thanks,
> Marcia
>
> Marcia Hart
> GRA
> Division of Biology
> Kansas State University
> 19 Ackert Hall
> Manhattan, KS
>
> Office:(785)532-6795
> E-mail: mlh2207@ksu.edu
>




PD Dr. Andreas Simm
Universitaet Halle Wittenberg
Klinik fuer Herz- und Thoraxchirurgie
Ernst-Grube Str. 40
D-06120 Halle

Tel.: +49 (0) 345 557 2647
		  552 2878
FAX:  +49 (0) 345 557 2782
		  552 2890



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:08 EST