Re: cell cycle analysis with GFP and Hoechst not working

From: Howard Shapiro (hms@shapirolab.com)
Date: Sat Apr 20 2002 - 08:04:59 EST


Ann Atzberger wrote-

>have had same problem with drosophila. Solution is simple. Put a density
>filter in front of the GFP detector. This will correct the "Hoechst
>contaminating fluorescence" leaving your GFP signal intact. The problem is
>not the detector but the sample itself  (but not all samples are like that-
>it might be the protocol).
>
>Also when aligning the instrument make sure you use 488nm beads and UV
>beads which only fluorescence in their specific channels -
>
>If you have the same configuration as I have the slot for the density
>filter is before all the pmts off the 1st laser -so you will have to
>increase the SSC signal.
>
>Once you get this working you'll find it's a very nice assay.
>
>hope this helps- if you come up with another solution let me know.

In response to Renee Read's:

>>      I am a graduate student and I've been using flow cytometry to study
>>the cell cycle in Drosophila cells.  I dissociate cells from tissue
>>samples and stain for DNA content with Hoechst 33342 and it works
>>great.  However, I've been trying to also use GFP as a marker in these
>>assays and I've found that the Hoechst flouresces in the GFP channel and
>>obscures the GFP signal to the point where compensation can't fix the
>>problem.  We are using a Cytomation Moflo mahcine and we are using the
>>UV laser for the Hoechst and the FL2 detector for the GFP.  We've tried
>>shutting off the UV laser and the Hoechst background is still picked up
>>in the GFP chanel, indicating the problem is with the FL2 detector.  The
>>flow technician isn't sure how to fix the problem.  Do you guys have any
>>suggestions on how to look at cell cycle and GFP flourescence in live
>>cells?

This is not making a great deal of sense to me. If there is some spectral
overlap between Hoechst fluorescence and GFP fluorescence, a neutral
density filter will do nothing to improve separation; if the problem is
that the green fluorescence signal is too high, then lowering the PMT
voltage instead of putting in an N.D. filter should fix the problem  -
unless Ann is using the N.D. filter because the PMT voltage is already as
low as is recommended.

But the other question is: what fluorescence from Hoechst is being excited
at 488 nm? Interference with fluorescein (and PE) fluorescence is fairly
common in formaldehyde-fixed cells stained with Hoechst, but appears to be
due to some reaction between the fixative and the Hoechst dye. That is not
what is going on here. Hoechst and DAPI do form some yellow fluorescent
products with materials other than DNA (and I'm not sure anybody knows what
they are) in some bacteria (and, if I remember correctly, also some plant
cells). My point is that you ought to get a handle on where the interfering
fluorescence is coming rather than trying to correct for it by brute force.

-Howard