RE: Fc-Blocking

From: Fischer, Randy (NIAMS) (fischer1@mail.nih.gov)
Date: Fri Apr 12 2002 - 08:58:00 EST


Volker,

The answers to your questions are not as simple as you might like.

1.  Blocking: Depends.  We have found no significant difference using mouse
IgG1 antibodies to stain human PBMC with or without blocking.  When we use
mouse IgG2a/b, we must block with rat serum or we get higher backgrounds.
And when we do staining that requires secondary antibodies that are goat or
rabbit (yes, some of our antibodies still require secondaries), we find it
necessary to block with those sera as well.

2.  How fast/how long: Again, I am sure my opinion is one of many different
ones you will encounter.  But, we stain 100,000 cells and I routinely
recover at least 80% after multiple washes and fix/perm procedures.  Our
speed/force is 450xg for 10 minutes if using 15-50mL tubes.  For the 12x75mm
tubes we spin at 450xg for 5-6 minutes.

3.  Buffers:  Well, here is that black box of cell biology.  We all know
from years of experience/dogma that cells require protein to keep them
"happy".  However, I have used straight DPBS without Ca/Mg for staining
human PBMC and the cells did fine, but it was a quickly done test that
required minimal time.  So, we routinely use the DPBS with 1% BSA to provide
a little extra buffering to keep the cells happy in case we have to leave
them a little longer than expected, or if it is one of our multiple step
staining procedures.

Good luck,

Randy T. Fischer
NIH/NIAMS
Building 10, Room 6D57
9000 Rockville Pike
Bethesda, MD 20892
(301) 594-3537
fischer1@mail.nih.gov

> ----------
> From:		Eckstein, Volker
> Sent:		Thursday, April 11, 2002 7:28 AM
> To:	Cytometry Mailing List
> Subject:	Fc-Blocking
>
>
> Hi flow-experts,
>
> Fc-blocking (or not) on human PBMC prior direct or indirect staining is a
> never ending discussion or story in our lab.
> And the literature is confusing.
> Some claim to use an IgG solution of the same origin as the used cells
> are.
> Others an IgG solution of the same origin as the monAb (direct coupled
> with
> fluorochromes) or the secondary polyclonal Ab are. What is the flow
> community thinking about that issue and is there really a need?
>
> Some more questions:
> What is the best wash and staining buffer for human PBMC staining
> protocols?
> Some people add BSA or HSA and some FCS in different concentrations (0.1 -
> 5%) to a PBS-buffer.
> And what is the cell biological need to add serum components? I suppose
> not
> blocking Fc-receptors. The explanation I've red that cells feel at home
> and
> comfortable is unsufficient (for me).
>
> And finally the question of speed in centrifugation of the cells for
> washing
> after staining.
> Mostly published protocols differ extremely. The range in speed starts
> with
> 200g up to 1800g or more for 1 to 10 minutes. It depends on the source of
> cells that's clear but even for human PBMC there are such differences
> outlined. PBMC's are hard taking cells but where is the balance between
> cell
> loss by damage (high g) or cells loss by aspirating due to less pelleting
> of
> the cells (low g)?
>
> All responses are highly welcome.
>
> Best regards
>
> VolkerVolker Eckstein PhD
> Dept. of Internal Medicine V
> Medical School of the University
> University of Heidelberg
> GERMANY
> volker_eckstein@med.uni-heidelberg.de
>
>
>
>
>
>