If you co-layer the UV and the HeNe in the 3rd position you can detect the red signal from the HO dye in FL3. My LSR is configured with a change in laser alignment and pinholes to allow UV/HeNe laser to be seen in FL3. Not so hard to do. Contact BD if unable to re-align yourself. Jim Houston > ---------- > From: David Dombkowski > Sent: Monday, March 4, 2002 7:32 AM > To: Cytometry Mailing List > Subject: SP with LSR > > > Kimmo, > > You will not see red fluorescence with Hoechst 33342 with the > configuration of 2 PMT UV, 3 or 4 PMT Argon and 1 PMT He-Ne. The red > fluorescence is too high for dichroic to reflect to PMT for UV. You need > to > be able to collect this signal in PMT usually dedicated to Argon or He-Ne > laser. There is no pinhole for UV fluorescence in these channels. You need > a custon configuration which may be achieved if you cantact Becton > Dickinson. > > David > > David M. Dombkowski > dombkowski@helix.mgh.harvard.edu > Flow Cytometry-Pathology-CNY rm7017 > 149 13th Street > Massachusetts General Hospital-East > Charlestown, MA 02129 > Tel. (617)726-1683 > Fax.(617)724-3164 > > > >