different count in immunocytochemistry and FACS

From: Dr Mark W Lowdell (m.lowdell@rfc.ucl.ac.uk)
Date: Tue Feb 05 2002 - 05:40:39 EST


Rosetta - this is almost certainly due to the mathematics of rare-event
analysis.  The distribution of rare events follow Poisson statistics and
the reproducibility of your estimates of "positive" events is determined
by the number of positive events scored.  You don't say how large your
sample size was in each of the examples given but I would guess that the
immunocytochemistry data are based upon a few thousand cells at best.
Using Poisson statistics we know that the c.v. of the estimate of a
reare event is given by the following formula -
Cv = 100/ sqrt "n" (where "n" = the number of positive events

E.g  acquire a data file with 100,000 total events of which 100 are
positive for your antigen of interest.	This gives a frequency of 0.1%
and the coefficient of variance of this result is 10%.	Once you get
below 100 positive events the c.v. increases exponentially and your
precision decreases similarly.

Dr Mark W Lowdell
Senior Lecturer in Haematology /
Head of Laboratory of Cellular Therapeutics
RFUCMS

T - 020 7830 2183
F - 020 7794 0645