Re: 7-AAD/PE problem

From: DAVID M CODER (d_coder@MSN.com)
Date: Thu Jan 31 2002 - 19:18:12 EST


I agree that the apparent reduction of FITC signal (and not PE) for
cytoplasmic proteins is not at all clear. (Were the experiments repeated
using only TO-PRO-3 and a red laser where PI crosstalk is eliminated?) For
nuclear antigens in close proximity to DNA, energy transfer would be
expected for the cases note below. p53 can exist in both cytoplasmic and
nuclear compartments, so one would expect differences related to the
position of a cell in the cell cycle.

Dave

----------------
David M. Coder, Ph.D.
Consultant in Cytometry
email: d_coder@msn.com
tel./messages: 206-499-3446

----- Original Message -----
From: "Corver, W.E. (PATH)" <W.E.Corver@lumc.nl>
To: cyto-inbox
Sent: Thursday, January 31, 2002 6:14 AM
Subject: RE: 7-AAD/PE problem


> ----- Original Message -----
> From: "Corver, W.E. (PATH)" <W.E.Corver@lumc.nl>
> To: cyto-inbox
> Sent: Tuesday, January 29, 2002 9:06 AM
> Subject: RE: 7-AAD/PE problem
>
> David and Brian wrote:
>
> >
> Energy transfer seems the best explanation. FITC and PI are a near optimal
> pair; PE and PI less so, but still there is overlap. PI and TO-PRO-3 are a
> very good pair: the emission of PI overlaps well the absorption of
TO-PRO-3
> (e.g., a HeNe laser at 633nm excites TP-3 very well), and both dyes
> intercalate into DS-DNA. In fact, the only way to observe TO-PRO-3
> fluorescence when only a 488nm line is available is by transfer from PI.
All
> such energy transfer (FITC, PE --> PI; PI-->TP3) is avoided with direct
> excitation of TP-3 by red (633 to 647nm) excitation.
>
> Best regards,
> Dave
> ----------------
> David M. Coder, Ph.D.
> Consultant in Cytometry
> email: d_coder@msn.com
> tel./messages: 206-499-3446
>
>
>
>
> > -----Original Message-----
> > From: Newsom, Brian S. [mailto:BSNEWSOM@txccc.org]
> > Sent: maandag 28 januari 2002 22:02
> > To: cyto-inbox
> > Subject: 7-AAD/PE problem
> >
> >
> >
> > Question to all of you DNA experts. We have someone who is tring to run
an
> > experiment staining a nuclear localized protien with PE and then
staining
> > with 7-AAD. The 7-AAD staining works well with a good CV. The antibody
> > staining looks decent (although low percentage and fairly dim) when by
> > itself, but when the 7-AAD is added the antibody staining totally goes
> away.
> > Is there an energy transfer or steric hinderance issue that may be going
> on?
> > Any help appreciated.
> >
> > Brian
> >
> Dear David,
> Indeed PI and TO-PRO-3 are excellent candidates for FRET, as we and
several
> others have published. Also, FRET might be involved in this case. However,
> we are still puzzled. We observed a significant reduction in cell surface
> fluorescence (FITC and PE) after adding PI at high concentrations (50
µg/ml)
> using viable cells. Using TO-PRO-3 (lower concentrations, 2 µM) we
obtained
> comparable results. We could not explain this quenching of fluorescence
due
> to absorption (unlikely due to the law of Lambert-Behr) nor FRET (we did
> some measurements). Especially the effect of TO-PRO-3 on FITC is
intriguing
> due to large differences in exc./emis. wavelengths of the dyes. Maybe
Howard
> has an idea.
>
> Willem E. Corver, PhD
> Department of Pathology
> Leiden University Medical Centre
> P.O. Box 9600, Building 1, L1-Q
> 2300 RC Leiden
> The Netherlands
> Tel: +31 71 5266604, Fax: +31 71 5248158
> E-mail: W.E.Corver@lumc.nl



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:25:58 EST