Re: dye laser details

From: Howard Shapiro (hms@shapirolab.com)
Date: Sun Jan 20 2002 - 11:20:41 EST


Rachel Gerstein wrote-

>Could someone explain how the tuneable part of the dye laser works ? What is
>being altered such that different wavelengths result ?

To make a usable laser, one puts mirrors around a medium in which
stimulated emission is occurring.  The rear mirror, often called a "high
reflector", is very highly reflective at the lasing wavelength(s); the
front  mirror is also, as mirrors go, highly reflective, but lets a small
fraction of the light at the lasing wavelengths pass through, and is thus
usually called an "output coupler".

In a lasing medium such as ionized argon, there are only a certain number
of fluorescence transitions at which lasing can occur; this explains the
discrete lines which can be obtained from an argon ion laser.  If the
mirrors are coated to be reflective across a broad spectral range, it is
possible to obtain simultaneous emission of several spectral lines in the
visible and/or ultraviolet (the "and" in the "and/or" here covers the cased
of a laser such as the Coherent Enterprise, which can be configured for
simultaneous UV and 488 nm output).

If one inserts a spectral dispersal element - the best known being a
Littrow prism - between the lasing medium and the high reflector, different
wavelengths of light will be diverted at different angles, and the range of
wavelengths that can strike the high reflector at the near-normal incidence
required for the reflected light to be directed back through the medium
(and then to the output coupler, back through the medium and the prism to
the high reflector, etc., maintaining laser action) is sharply restricted.
Thus, in an argon laser, the prism allows one laser line at a time to be
selected; which line it is depends on the angle at which the prism is set.

In some cases, even discrete laser lines "compete" with one another (for
example, competition between yellow and red lines in krypton lasers is
fairly noticeable) and it is necessary to use either a prism or mirrors
with reflectivity restricted to a narrow spectral range to get optimal
output of one of a number of competing lines.

In dye laser, the lasing medium is a fluorescent organic dye.  The
fluorescence emission spectrum of a dye is broad, because the polyatomic
structure allows some of the excitation energy to be lost by one or more of
many possible nonradiative transitions before fluorescence emission occurs.
Laser action can occur across a fairly broad region of the emission
spectrum, but the different wavelengths are effectively competing with one
another.  In order to get practical amounts of light out of the system, it
is necessary to restrict the output wavelength to a much narrower range;
this is accomplished using a Littrow prism almost exactly as one would be
used in an ion laser.  However, since the dye emission spectrum is
continuous, one gets a continuously tunable laser output from the dye
laser, whereas, with the ion laser,  laser action occurs only at the
discrete lines characteristic of the gas medium.

-Howard



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:25:56 EST