discordant results from flow and microscopy

From: Ann Atzberger (ann.atzberger@EMBL-Heidelberg.de)
Date: Mon Jan 21 2002 - 04:42:16 EST


Hi Lynn,

I've had a similar experience but a different kind of assay. We were doing
phagocytosis assays with Mosquito cells and fluorescent bacteria. The
results on the microscope were very different to what we were getting on
the Facscan-  Anyway what was happening was that these cells had loads of
vacuoles  which were also fluorescing in the green channel. You could see
this on the microscope- and in the end we gave up on the FC.  You might
have a similar problem - so by microscopy you need to see if other
components of the cells are also fluorescing e.g yellowish, orangy,
greenish etc.

I would also suggest decreasing the p-formaldehyde concentration if
possible as it may also contribute to "unwanted fluorescence".


Switching to Fitc as a marker might be an option as from personal
experience I find Alexa 488 to be very bright (for HeLa- PMT 427) and as
you say the cytosolic antigen expression is a punctate distribution- maybe
the PE signal is simply too weak when used in conjunction with a very
bright marker?? Could that happen, flowers??


Have you compared -by microscopy- the sample prepared for FC and the sample
prepared for microscopy?


can't think of anything else
regards
Ann