Re: HO33342 into 670/20 filter???

From: Larry Arnold (lwarma@med.unc.edu)
Date: Fri Jan 11 2002 - 16:12:20 EST


Maris

What instrument are you using?  HO33342 does emit significantly at that
wavelength so you can get cross-talk between lasers.  After I know what
instrument you are using I may be able to suggest something.

Larry

At 11:12 AM 1/10/2002, you wrote:

>Hi Everyone,
>Happy New Year to All!
>
>I am at a loss to explain why cells that are stained with Hoechst33342 are
>increasing
>the background in my APC signal...HELP!  We were looking at DNA using
>HO33342 (excited by
>UV, emission through a 450/20 filter) and APC (surface receptor, using
>647nm excitation
>and emission through a 670/20 filter).  There was nothing out of the
>ordinary visible
>during the alignment (no laser noise increasing the background of either
>signal, etc).
>We set the APC negative control within the first decade, looked at the
>APC+ control, and
>saw a distinct positive population.  When we put on a sample that was
>stained with HO
>(APC-), the signal in the APC channel went into the higher end of the
>second decade.
>At first, we were splitting the two signals with a 510LP, and after
>witnessing the
>shift in background, we switched to a 640LP...we still saw the same shift.
>We compensated the HO out of the APC signal, and the experiment seemed to
>work as
>expected (read: the researcher was happy with his results), but does
>anyone have any
>idea what happened?  I've been looking at every spectrum figure I can
>find, and have
>looked at the filter specs and I can't find any reason for that
>shift.  Any advice,
>as always, is sincerely appreciated. Thanks,
>Maris
>
>Maris A. Handley
>Dana-Farber Cancer Institute
>J415
>44 Binney St.
>Boston, MA 02115
>(617)632-3139
>maris_handley@dfci.harvard.edu

Larry W. Arnold, Ph.D.
Associate Professor
Director, Flow Cytometry Facility
Department of Microbiology and Immunology
Lineberger Comprehensive Cancer Center
CB# 7290
University of North Carolina at Chapel Hill
Chapel Hill, NC 27599
Phone: 919-966-1530
FAX: 919-962-8103



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:25:55 EST