---------------------- Forwarded by Ian DIMMICK/Europe on 09/01/2002 17:29 --------------------------- Ian DIMMICK 08/01/2002 20:14 To: cyto-inbox cc: Subject: Re: meaning of FL2-area and FL-2 width (Document link: Ian DIMMICK) Dear Janet, (Embedded image moved to file: pic00041.pcx)(Embedded image moved to file: pic18467.pcx) >From the above two diagrams you can see the generation of a pulse in diagram 1 as a cell passes the interrogation point where the laser is focussed on your cell nuclei, in the second diagram the pulse can be processed in various ways The height of the pulse is proportional to the intensity of the fluorescence signal (in your case PI) and in part due to the shape of the laser interrogation spot. The width of the pulse is proportional to the time taken for the cell nuclei to traverse the interrogation point of the laser (so called time of flight) The area of the pulse is calculates from the integral fluorescence (area) that is cross hatched in diagram 2 The reason for using these two parameters(area v width) is primarily to discriminate single cells passing through the flow cell from cells that may have no spacial separation when passing through the flow cell ie doublets.Doublets will give an increased width measurement (26% increase for every doubling of volume of a cell based on a spherical arithmetic model), The area measurement will also increase however when analysed the histogram of width v Area will enable you to gate out any doublets(disproportionate width to Area) and therefore avoid mis interpereting a Go doublet for a G2m event Hope this helps Ian
This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:25:55 EST