Re: DNA Measurement

From: Derek Davies (daviesd2@icrf.icnet.uk)
Date: Fri Dec 14 2001 - 09:25:16 EST


Hi Ron,

I have found that this isnt an uncommon phenomenon but one that is often
overlooked. Assuming you are using pulse processing to exclude the
obvious doublets quite often in established cell lines there is a
re-duplicated population containing '8n' cells. Cells that are
artifically blocked in G2 by, for example, nocodazole do sometimes show
cells that apparently double their DNA but do not divide. If there is
excessive clumping (and pulse processing will not be able to remove 100%
of clumped cells), you will see peaks at the 6n, 8n, 10n and so on.
Generally in reduplication, the 8n peak clearly looks like a 'real'
population. Its always worth saving your data ungated and checking
during analysis.

We have seen in certain cell types that transfection can lead to changes
in the DNA content of cells. Maybe by adding an internal control of
known DNA content you can re-assure yourself that you are actually
seeing an increase in DNA content rather than it being a staining
artefact (although to see such a large change due to something like
changes in DNA conformation might be unusual in itself). Sorting the 8n
population (or using a Laser Scanning Cytometer) will allow you to see
whether they are single large cells or clumps.

Hope that helps,
Derek


On Tue, 11 Dec 2001, Ron Hill wrote:
>  One of our users has observed an apparent doubling in the amount of
>  DNA (based on PI staining) in the 2N population of NIH3T3 cells after
>  selection in drug to establish cell lines that stably express a gene
>  of interest. The cells were engineered by retrovirus. The doubling is
>  compared to the parental cells. The selected cells have a 4N
>  population with 4 times the DNA of parental 2N cells. Has anyone ever
>  seen this. Is it likely to be some artifact of staining?

************************************************************************
Derek Davies                       Voice: (44) 020 7269 3394
FACS Laboratory,                   FAX: (44) 020 7269 3100
Imperial Cancer Research Fund,     e_mail: derek.davies@icrf.icnet.uk
London, UK			   mobile: 07790 604112

Web Page: http://www.icnet.uk/axp/facs/davies/index.html

In tenebris lux
*************************************************************************



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:43 EST