RE: cd4 cd8 coexpression

From: Alice L. Givan (Alice.L.Givan@dartmouth.edu)
Date: Tue Nov 06 2001 - 19:45:17 EST


Hello Flowers,
I just wanted to re-inforce Ken Ault's comments about artifactual co-expression of CD4
and CD8 due to coincidence of two cells in the laser beam.  Two cells can coincide in
the beam either because they are physically aggregated into a clump or because they
just happen (by statistical probablility)  to be suspended in the same volume of sample
buffer as it moves past the laser.

 A coincidence artifact should be suspected if:
1) as Ken said, the frequency of these CD4/CD8 doubles decreases when the flow rate
is decreased (although this may not happen if the cells are in actual clumps).
2) the intensity of each color on the double expressors  is  the same as the intensity of
each color on the relevant single expressors.  In other words, if the double expressors
form the fourth corner of a perfect rectangle on a dot plot (with the negs, the PE+
singles, and the FITC+ singles forming the other three corners),  then you should
be suspicious.

Alice


Alice L. Givan
Englert Cell Analysis Laboratory
of the Norris Cotton Cancer Center
Dartmouth Medical School
Lebanon, New Hampshire NH 03756
tel 603-650-7661
fax 603-650-6130
givan@dartmouth.edu



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:38 EST