CD4+/CD8+/CD3+ cells are 1-4% range of the lymphocyte population from our HIV+ patients. Most of the cells are brightCD4+/dimCD8+ but all combinations of bright and dim are possible. We include these dual positive cells in both the CD3+/CD4+ and CD3+/CD8+ counts. What do other labs do with these cells and why? -----Original Message----- From: Kenneth Ault [mailto:aultk@mmc.org] Sent: Wednesday, October 31, 2001 7:32 PM To: cyto-inbox Subject: Re: cd4 cd8 coexpression A phenomenon frequently forgotten is coincidence. If two cells enter the observation volume at the same time they will be seen as one event with the properties of both cells. This is a very common problem in my world (platelets) and should be considered as a possible explanation for any kind of unexpected dual expression of markers. It would be interesting to know if the frequency of CD4/CD8 doubles changes as the particle flow rate changes (i.e change the sample pressure or dilute the specimen.) Ken Ault