Hi Everyone, If you have a 0-10 PSI pressure gauge with a set of quick disconnect fittings, measure the sheath tank pressure. It should be 4.5 PSI. If it is not 4.5 PSI adjust the regulator, located in the lower right portion of the instrument, for 4.5 PSI. Once 4.5 PSI sheath tank pressure has been established, measure the sample tube pressure. This can be measured at green line going into the flow cell. A ball park measurement should be 4.1 PSI for low, and 4.6 PSI for high. Sheath fluid pressure will be lower at your flow cell than in your sheath tank. Your flow rate in high should be 60uL/min and in low, 12uL/min. If you would like a pressure gauge contact me. Hope this helps. Ray Raymond Lannigan Tritech Inc. 2014 Renard Ct. Suite I Annapolis, MD 21401 1-410-266-1522 1-800-886-7004 www.tritechinc.com -----Original Message----- From: Marco A. Fernández <marcoa@ns.hugtip.scs.es> To: cyto-inbox Date: Friday, October 26, 2001 3:42 PM Subject: Sample pressure > >Dear Flow'ers, > >We have started a discussion in our lab regarding sample pressure vs. sheath >pressure in FACS benchtops (we have a FACScan). I am not sure but I think >that sample pressure is always over sheath tank pressure (4.5 psi for sheath >and 4.6 -LO- and 5.0 -HI- for sample). Is that right? > >Best regards, > > >Marco A. Fernández >FACS Facility >University Hospital Germans Trias i Pujol >08916 Badalona (Barcelona) >Spain >marcoa@ns.hugtip.scs.es >http://citometria.furebtip.org >